Can anyone suggest the feasible way to measure humidity and temperature at high temperatures (say 800deg C)? Most of the sensors and loggers available in market works up to 80deg C or120 deg C.

Where can i find such instrument????
 
C

Curt Wuollet

It isn't really humidity above 100C. It's percent water vapor. You might try searching for that.

Regards
cww
 
M

Marcos T. B.

This kind of measurement is used in drying chambers to optimize energy consumption.
They use zirconia probes (the same used in BMS systems), which actually measure %O2.
“Standard atmosphere” values of O2 indicate low moisture content. Low values of O2 indicate high percentage of vapours.

Contact Yokogawa or Pleva.

Regards

Marcos T. B.
 
R

Richard Pearson

We have an instrument that we check our drying hoods with that is rated to 650 deg C and can be retrofitted for higher temperatures.

it is by MAC Instruments
there are various models based on the % moisture by volume that you want to measure.

We have the MAC125-60

Web site http://www.macinstruments.com
phone 419-621-2322
 