Acika,

What is the purpose of this system? Is it just for keeping track of the calibration equipment? Or for the devices which the calibration equipment will be used for checking/adjusting?

Is this for companies providing calibration services?
 
The application can be used as a record of all measuring equipment. Or in brief outlines by areas:
  1. MSA - vision of the measurements obtained from the analysis of the measurement system / process, when they were made, whether they meet the VDA requirements, .
  2. Test equipment record as a card with all the data for the measuring device, such as date of purchase, type of device, serial number, accuracy, resolution, ...
  3. Certificates - Records of certificates obtained for each measuring devices.
  4. Acredited laboratories- all laboratories involved in calibration of measuring devices
  5. Standards - representation of the standards used in calibration of measuring devices - Self calibration-the values from the measurements from the calibrators, and gave and fulfilled the tolerances, the uncertainties, ...
  6. List of self-calibrators - records of all types measuring devices, for which we performed self-calibration. Which etiquette do we use, which measurement method, ...

A short presentation you can find on next link
 
So, when I think of "calibration" I think of the devices in a process plant which have to be calibrated prior to installation, or verified periodically, using regularly certified calibration equipument ("testers," or multi-function calibrators, etc.). As examples: pressure transmitters, temperature transmitters, temperature switches, pressure switches, etc. For this most process plants use some kind of "calibration data sheet" which lists manufacturer's information, device tag/number, calibration data (range; deadband/hysteresis; actuation/de-actuation setpoints; etc.), as-found condition, as-left condition, equipment serial number, and the like. These are usually stored in a location where they can be quickly accessed; I would imagine some sites have some kind of computerized calibration record management system where information is stored for future reference or review.

I don't think most process plants do anything more with their calibration equipment than send it out regularly for testing and certification by a company that issues certificates of accuracy and compliance. (If they even do that. A LOT of plants these days outsource device calibration and testing and repair.)

So, perhaps we have both learned something here. Calibration can refer to instrument calibration/testing/verification, and it can refer to the equipment used to calibrate/test/verify instruments.
 
Thank you for the suggestions. The idea of the application is more towards recording the measuring equipment but not recording the measuring results. If you want, I can share with you a version of the application, so you can make an review and a suggestion for improvement.
 